Since the first transesophageal echocardiography examination in 1975,
using M-mode imaging with mechanical scanner contained in oil bag having
limited control of direction, there have been a lot of modifications and
technical advances in the equipment1. In its infancy, the main
application of TEE was assessment of global left ventricular function
and regional wall motion abnormalities (RWMA). In the modern equipment,
with provision of multiplaner & multifrequency probes (3.5 to7 MHz),
colour flow mapping, pulse & continuous wave doppler, cine loop
displays and digital image processing, now it has become possible to
have same information rather of better quality than transthoracic
echocardiography. Another advantages of TEE include the stability of the
transducer and continuous recordings for extended periods of time, which
are very desirable for monitoring in operation theatre. With advancing
technology and experience of anaesthesiologists, the number of clinical
applications and scope of TEE is increasing further and getting more
defined.
Indications The common indications are:
1. Assessment of valvular function intra and postoperatively.
2. Evaluation of mitral valve repair and prosthetic valve surgery.
3. Assessment of global left ventricular function and regional wall
motion abnormalities (RWMA).
4. Assessment of aorta for arteriosclerosis, calcification and
dissections.
5. Detection of intracardiac defects, masses and vegetations.
6. To see optimal deairing after open-heart surgery.
7. For quick evaluation of severe hypotension intra and postoperatively
by ruling out conditions requiring immediate intervention like cardiac
temponade,poor left ventricular volume status ,RWMA , mal function of
prosthetic valve or systolic anterior motion of septum after MVR and
aortic dissections,etc.
8. Inadequate or impossible transthoracic echocardiography. Other areas
where its use is increasing these days are critical care units and
emergency rooms to have quick evaluation of polytrauma and sick patients
to ascertain the cause of shock.
Although TEE is a fast and relatively non-invasive technique, its
routine use for noncardiac surgery is yet to be established. Many
studies have revealed that echocardiography is more sensitive than
electrocardiography for early detection of ischaemia2-6. But as it is
difficult to monitor all areas continuously, it gives low yield even in
coronary artery disease patients undergoing noncardiac surgery. Secondly
as the probe is placed after induction of anaesthesia and removed before
extubation ,critical periods of stress are left unmonitored.However
quick assessment of loading status and measurement of cardiac output can
be very useful in variety of surgeries involving major shifts in the
intravascular volume of the patient. Another indication of TEE in
noncardiac anesthesia is to detect persistent fossa ovalis and venous
air embolism in intra cranial and orthopedic surgery.
Equipment and Its placement
TEE equipment includes the basic echocardiography unit and TEE probe.
All TEE probes have some common features. Currently employed equipment
hoses a 5 MHz transducer in the tip of probe. The tip can be directed
antero-posteriorly and sideways by adjusting the two knobs placed in the
proximal handle. Another feature is a temperature sensor at the tip to
warn of possible heat injury. Often TEE probe is placed once general
anaesthesia is induced, however it can be placed in awake patients under
local anaesthesia (Xylocaine 2% viscous and Xylocaine 4% topical spray)
and mild sedation with midazolam 2 mg . It is always safe to use bite
block in awake patients. Probe should be lubricated with xylocaine 2%
jelly and introduced gently through oral cavity till 25 cm mark at
incisior level, with transducer facing anteriorly. At this level one may
see 3 or 4-chamber view and can adjust gains and depth.
Standard Projections
Once TEE probe passes 3-5 cms down to inferior constrictor (20-25 cms
from incisors), one starts getting images of great vessels and cardiac
chambers. The common projections of interest to an anesthesiologist are:
Transverse Views
- Basal View or 3Chamber view In transverse axis at about 25 -30 cms
from incisors this view is obtained. One can see aortic valve and
sinuses in the center,coronaries,parts of both atria and interatrial
septum. This view can be used to study aortic valve anatomy, degree
of calcification and measuring aortic cross sectional area.Adjusting
the probe little in and out one can focus the pulmonary artery and
its main branches(Fig 1):
- Mid esophageal View or 4 Chamber View ( Fig-2).:
Once probe is further enhanced 2-3 cms down(30-35 cms from incisiors),4
chambers are seen in a single viewThis provides plenimetry of
chamber sizes,detection of intracardiac air bubbles after open
cardiac surgery and mitral valve evaluation.Detailed mitral valve
examination can be done using different angles to study mitral
leaflets, their coaptation and regurgitation in systole,mital valve
flow studies using pulse wave Doppler or continous wave Doppler and
colour flow mapping.This view is very useful judging the adequacy of
repair of mitral valve.
- Short-axis Transgastric View (Fig 3):
This view is obtained by introducing further to 40 cms or so and
anteflexing the probe so that transducer faces towards the apex of
heart.This view is very useful for monitoring LV function,studying
RWMA,ischaemia,LV aneurysm and preload of LV.
Longitudinal Views(90 degrees)
- Wrap View or RVOT View. (Fig 4):
At basal level once 3 chamber view is obtained,the probe is rotated
to 60 to 90 degree angles to get the Right atrium,RV and pulmonary
trunk in one view.At times one can see the PA catheter transversing
to one pulmonary arteries at this level.
- Midoesophageal View
This view is obtained by rotating the probe anticlockwise at mid
esophageal level and is used to make the detailed evaluation of left
atrium , its appendage(LAA),left pulmonary veins and pulmonary
venous flow pattern.
- Trans gastric View
In longitudinal view it is possible to see longitudinal section of
LV,papillary muscles and LA.
Common Clinical Applications
- Left Ventricular assessment
Global LV assessment includes measurement of dynamic indices and
segmental wall motion abnormalities. Assessment of global and
regional ventricular performance has become the cornerstone for
evaluation of patients with ischemic heart disease The ejection
fraction (EF),a commonly used contractility index can be derived
from the ventricular volumes in end -daiastole(EDV) and end systole
(ESV) using the following formula:
EF % = EDV-ESV/ EDV x 100
A fractional area change can be calculated from the 2D trans
esophageal echocardiography images,end systolic area ,and end
daiastolic area obtained from the transgastric short axis view at
the midpapillary level.
FAC% =EDA-ESA/EDA x100
Fractional shortening is inaccurate if ventricular asynergy
exists. Further it has been shown that EF is more sensitive than
cardiac out put measurement in detection of cardiac contractility8
to decrease further ambiguity from dyskinesia, it is possible to
measure LV volume using Simpson's formula,although it is impractical
to use it in routine clinical settings9. Measurement of
wall thickening and assessment RWMA are important tools to assess
cardiac contractility.
Diastolic dysfunction can be checked using mitral valve flow studies
with pulse wave Doppler or continuous wave Doppler. Normally E wave
(wave of early ventricular filling) is greater than A wave ( wave of
late ventricular filling by atrial contraction ).In diastolic
dysfunction A wave is greater than E.
- Detection of Myocardial Ischaemia2-7,13
The superiority of TEE in the detection of
myocardial ischaemia over traditional monitoring like EKG and
pressure data changes has been proved by various authors3-7.
Transesophageal echocardiography clearly delineates regional wall
motion abnormalities. The changes in global systolic function &
diastolic dysfunction do not match the efficacy of RWMA to detect
the onset of ischaemia. The onset of visible RWMA within 10 to 15
seconds of coronary occulusion has been demonstrated experimentally3-4
The trans gastric short axis view of the left
ventricle is obtained at six levels to evaluate different parts of
LV for RWMA. Echocardiographic criteria for RWMA include an
alteration in left ventricular radial shortening and wall thickening
that usually accompanies normal contraction6. Normal
contraction is defined as greater than 30% shortening of radius from
the centre to the endocardial border. Mild hypokinesia relates to
shortening of radius between 10 % to 30%. Severe hypokinesia is
defined as less than 10% shortening, while akinesia is the absence
of motion. Dyskinesis is paradoxic outward movement during systole7.
Segmental Wall Motion Scoring System
| Wall motion |
Radial Shortening |
Myocardial Thickening |
| Normal Motion |
less than 30% |
+++ |
| Hypokinesia |
10-30% |
+ |
| Akinesia |
0% |
0 |
| Dyskinesia |
systolic lengthening |
systolic thinning |
-
Assessment of Mitral valve and its repair:
After examination of the anatomic features of the valve, colour flow
mapping and Doppler studies should be used to evalauate the flow
pattern across the valve.All regugitant jets should be traced to
thieir origin as occasionally pulmonary venous in flow jets can be
mistaken for mitral regurgitation.After surgical repair of the
mitral valve,the surgeons often used to check valvular competence by
injectining saline into the LV.Czer and colleagues found that, this
saline filling is associated with 40% false positive results and 30%
false negative results10.The only definitive test for mitral
competence is echocardiography.
-
Assessment of septal closures and repairs of Congenital Heart
Disease.
TEE is useful to detect the residual ASD and VSD defects.By
injecting saline into RA through central line with a syringe can
reveal residual ASD if any by showing air bubble shadows in left
atrium.VSD defects can easily be seen using colour flow mapping.
-
Assessment of Air Removal after Open heart surgery
Air collects often at nondependent sites of heart and major
vessels;apex of LV,aorta near origin of right coronary artery and
pulmonary artery.
-
Assessment of Prosthetic Valves
Each of the mechanical and bioprosthetic valve has characteristic
forward transvalular flow patterns, a certain frequency of
anticipated valvular regurgitation.In general ,Starr-Edwards(caged
ball mechanical prostheses) and bioprosthetic valves normally
exhibit a small central color Doppler signal of mitral regugitaion
confined to early systole. In contrast,Bjork-Shiley(tilting disc
type)exhibit small peripheral jets throughout the systole Problems
with these valves can be obstruction, valvular and paravalvular
significant regugitaions or infective vegetations.
-
Evaluation for Hypotension
During cardiac and non cardiac major surgery, hypotension is a
common occurance.The cause may not be apparent from routine
haemodynamic parameters.TEE affords essential or additional
information of the cardiac events causing
hypotension.Anaesthesiologists often use 4 chamber view and
transgastric view at mid papillary level for quick evaluation of
hypotension in the immediate post operative period or critical care
areas.It has been very useful to detect hypovolumia,cardiac
temponade,valve dysfunction,poor contractiality of LV,dissection of
aorta.In hypovolumic states,both papillary muscles are seen as
kissing to each other in transgastric view at mid papillary level
and this view is very reliable as 80% of the stroke volume is
ejected by contraction LV at this level11.TEE offers a
superior index of preload compared with pulmonary artery occuluding
pressure12.
-
Evaluation of Aorta for Cannulations and Dissections
Tee is very useful to assess aorta for atheromatous
plaques,calcification and dissections etc.Anaesthesiologists can
guide the surgeon to find the proper site for aortic cannultion.
-
Detection of intracrdiac masses,thrombus and vegetations.
TEE is better than transthorcic echocardiography in evaluation of
intracardiac masses because of better quality images and resolution.
-
Measurement of Cardiac Out put2,7,12
Cardiac out put can be derived from either 2D echocardiography or
Doppler imaging techniques.The accuracy of 2D scanning is limited by
geometric assumptions that this method implies.
Cardiac Output ( CO ) = (EDV-ESV) x Heart Rate
Where EDV and ESV are end diastolic and systolic volumes
respectively.
Doppler technique applies determination of cross sectional area of
aortic orifice and velocity measurement at the same level.Cross
sectional area of aortic orifice can be measured from 3 chamber view
( Fig 1) using either formulae from equqtion 1 or 2, and calculating
velocity time integral ( VTI ) using pulse wave Doppler at orifice
level through transgastric view.. Then cardiac output can be
measured by multiplying stroke volume to heart rate.
Cross sectional area= 0.433 x L2
Equation -1
Or
p (D/2)² = 0.785 x D2
.. Equation-2
SV = CSA x VTI
.Equqtion-3
Cardiac Out Put = Stroke Volume x Heart Rate
where L=length of cusp
D=Diameter of orifice
SV= Stroke Volume
CSA= Cross Sectional Area
VTI= Velocity Time Integral
Complications:
TEE is very safe and atraumatic if used gently. However few cases of
esophageal tears and burns have been reported. To avoid such
complications the probe should be introduced with gentleness and
care.And when lying idle in the esophagous the probe should be locked
and equipment switched off to prevent the burns.Absolute
contraindications to TEE in intubated patients are esophageal
stricture,tumour,recent suture lines and diverticula.Relative
contraindications include hiatal hernia,esophageal varices,esophagitis
and unexplained upper gastrointestinal bleeding.
In coming era, I hope TEE will become a routine tool in emergency
room to assess the critically ill and trauma patients to have quick
information regarding the patient's ventricular performance, loading
status, physiological and mechanical compressions by pericardial
effusion or temponade, myocardial contusions, major vessel tears and
haematoma. Another area which needs consideration is training and
quality assurance for TEE. Although there are no specific guidelines for
quality assurance so far in India and equipment is available only in
limited centres, but proficiency can be achieved by dedicated training
like cardiac fellowships, reading scientific fundamentals of
echocardiography, TEE atlas, self assessment, image re-evaluation by
cardiologists and practice.
REFRENCES
- Hisanga K,Hisanga A,Nagata K, Yoshida S: A new transesophageal
real time 2D echocardiographic system using a flexible tube and its
application.Proc Jpn J Med Ultasonogr 32:43,1977
- Cardiac Anaesthesia: By J .A. Kaplan, Steven N. Konstadt,
D.L.Reich, 4th edition (1999),W. B.Saunders Company,Phildelphia
- Labovitz A,Lewen MK, Keim m,et al.:Evaluation of LV systolic and
diastolic dysfunction during transient ischemia produced by
angioplasty.J Am Coll Cardiol 10:748,1987
- Boman LK,Clemn MW,Cabin HS et al:Dynamics of early and late
ventricular filling determined by Doppler 2D echocardiography during
percutaneous transluminal coronary angioplasty.Am J Cardio
61:541,1988.
- Battler A,Froechicher VF ,Gallagher KP et al:Dissociation between
regional myocardial dysfunction and EKG changes during ischemia in
the conscious dogs.Circulation :62;735,1980
- Clinical Transesophageal Echocardiography-A Problem Oriented
Approach: By Yasu Oka; Steven N. Konstadt, 2nd edition
(1996),Lippincott-Raven Publishers Phildelphia.
- Madan Mohan Madali: Transesopheal echocardiography and the
anaaesthesiologist: CME 4th SACA Congress,1999.
- Roizen Beaupre et al: J Vasc Surg. 1; 300-305,1984
- Gert Poortmans,, Guido Schόpfer, Carl Roosens, Jan
Poelaert:Transesophageal echocardiographic evaluation of left
ventricular function.J Cardio thoracic and Vascular Anaesthesia Vol
14(5), 2000.
- Czer LSC, MaurerG,Intraoperative echocardiography in miral and
tricuspid valve repair.Echocardiography 7:305,1990
- Clements FM Harpole JEB, Quill T, et al: Estimation of left
ventricular volume and ejection fraction by 2D transesophageal
echoceardiography: Comparison of short axis imaging and simultaneous
radionuclide angiography. Br J Anaesth 64,331,1990
- Laver MB: Myocardial ischaemia ,Dilemma between information
available and information demand.Br Heart J :52,222,1983
|
|